Design Of Slip Stream For Cdu Overhead Ph Analyzer
#1
Posted 29 July 2017 - 12:39 AM
Can you help me do a design of slip stream of overhead sour water from accumulator for pH monitoring?
Initially, my design consist of a stream from SW pump discharge with pressurr regulator and RO to have constant flow to analyzer. The stream will go back at pump suction. I intentionally locate the analyzer at a higher point to avoid solid particles. I was also planning to put a mesh strainer.
I dont think this is the best design for the intent. I want to ensure that the pH analyzer will work and less maintenance is required.
Thanks for the help.
#2
Posted 30 July 2017 - 05:41 PM
Vjmromero,
It´s seems adequate, but I attached some recommendations of specialist. Do you have a sketch?.
Napo.
Attached Files
#3
Posted 01 August 2017 - 01:29 AM
Thank you for your response. Also, please see detail of the drawing. I was also checking if Pressure regulator valve alone will suffice the requirement
Pressure - [ ] in kPag
Temperature - ( ) in deg C
Flow - < > in m3/h act
Attached Files
Edited by vjmromero, 01 August 2017 - 01:32 AM.
#4
Posted 01 August 2017 - 07:01 PM
Vjmromero,
Your design is correct, also we have similar designs in ours Cooling Towers.
Napo.
Similar Topics
![]() Scrubber DesignStarted by Guest_nasss_* , 07 Aug 2024 |
|
![]() |
||
Refinery Lpg Deethanizer Column DesignStarted by Guest_Ilyes_* , 15 Feb 2025 |
|
![]() |
||
Alkaline Electrolytic Cell/stack Sizing/design For H2 ProductionStarted by Guest_BRS09_* , 13 Mar 2025 |
|
![]() |
||
![]() Heat Exchanger Network DesignStarted by Guest_Kakashi-01_* , 21 Feb 2025 |
|
![]() |
||
How To Remove The Condensate In A Wet Gas StreamStarted by Guest_ksagarzazu_* , 25 Feb 2025 |
|
![]() |